annual machine calibration

Posted: Mon Jan 06, 2020 8:07 am
by roberto flores
We do our daily calibrations with a qual sphere, but what needs to be done for annual calibration sticker?
We have a small plastic box labeled "Keyence Calibration Kit" which contains a 0.1000 ceramic pin and 0.8000 ceramic ring, but i'm not sure what these are for.
Do i need to schedule a tech to come in and calibrate my machine, or is this something that can be handled internally?
if internal, does someone have a procedure for this?

Thank you

Re: annual machine calibration

Posted: Mon Jan 06, 2020 10:44 am
by CrashN8
Annual calibration will certainly entail more than two artifacts you've described. Traditionally the annual calibration will entail multiple point calibration - i.e. length standard - along with some artifact for size / probing. What system are you trying to calibrate? I assume it's running CMM-Manager software since you've asked this question at CMM-Manager Forum.

Re: annual machine calibration

Posted: Mon Jan 06, 2020 1:14 pm
by roberto flores
Its a Helmel Phoenix DCC 216 x 142.
we mainly use it as a "manual" CMM, meaning we drive it with the jog box, but we have a couple DCC programs for more common parts.
We also have 5 Brown & sharpe and 1 Zeiss Contura that get their annual calibration by factory techs in the manner described by crash.

Re: annual machine calibration

Posted: Mon Jan 06, 2020 1:22 pm
by roberto flores
Update, we sent some interns into the record room (dusty storage area of boxes and whatnot) and found our service contact.
Crisis averted!