Head Calibration Procedure

Post Reply
jpo_031
Regular User
Regular User
Posts: 20
Joined: Tue Jan 10, 2017 5:06 pm

Head Calibration Procedure

Post by jpo_031 » Thu Aug 09, 2018 9:17 am

Hi there. I wonder if you can walk me thru to perform the Head Cal.
I'm trying to do a head calibration, the purpose is to verify the accuracy of the cmm due to a head crash during the past days, but when i go to UCC server I can see the option to calibrate the tips but i can't find anything for head cal.

jpo_031
Regular User
Regular User
Posts: 20
Joined: Tue Jan 10, 2017 5:06 pm

Re: Head Calibration Procedure

Post by jpo_031 » Thu Aug 09, 2018 12:03 pm

I Forgot to mention I'm using a PH-20 head TP-20 module ,with the 3.6 cmm mgr version.
Also have this other question besides trying to calibrate the head which I'm still trying to figure out, is it better to measure thin parts like sheet metal or small machined parts disabling the PH-20 and using only straight movements or whats your suggestion.

ameade022008
Power User
Power User
Posts: 76
Joined: Thu Oct 03, 2013 12:48 pm
Location: Tempe, AZ
Contact:

Re: Head Calibration Procedure

Post by ameade022008 » Thu Aug 09, 2018 12:28 pm

--->UCCassist-2, play the sequences via that, or use the "tool" toolbar to "create shortcut" of the sequences
ucc diag.PNG
UCC.PNG
You do not have the required permissions to view the files attached to this post.
Albert Meade
Quality Manager, Metalcraft Inc.
http://azmetalcraftinc.com/

jpo_031
Regular User
Regular User
Posts: 20
Joined: Tue Jan 10, 2017 5:06 pm

Re: Head Calibration Procedure

Post by jpo_031 » Mon Aug 13, 2018 12:16 pm

Thanks for the heads up.
I see is in the software, so I have to update to 3.7 since the one I'm using is 3.6..

ameade022008
Power User
Power User
Posts: 76
Joined: Thu Oct 03, 2013 12:48 pm
Location: Tempe, AZ
Contact:

Re: Head Calibration Procedure

Post by ameade022008 » Mon Aug 13, 2018 2:18 pm

I was doing all the stuff mention on 3.6. I dont think thats the issue
Albert Meade
Quality Manager, Metalcraft Inc.
http://azmetalcraftinc.com/

User avatar
RodGarrett
Nikon
Nikon
Posts: 8
Joined: Thu Feb 24, 2011 7:42 pm

Re: Head Calibration Procedure

Post by RodGarrett » Thu Sep 20, 2018 7:55 am

Hello Albert,

From UCC Server you can calibrate your head with the command: "RWReQualify(Head())"

Upon head calibration all probe assemblies will need to be re-calibrated.